Author Topic: Optimax II-S melted insulator plate!
Tommie Evans
Expert Film Handler

Posts: 116
From: Birmingham, West Midlands, UK
Registered: Sep 2004


 - posted 10-08-2004 01:54 AM      Profile for Tommie Evans   Email Tommie Evans   Send New Private Message       Edit/Delete Post 
Hi, I'm relatively new to the Strong Optimax II-S lamphouse.
Came in this morning to find the insulator plate (positive binding post) has atomised. Any suggestions on how this may have happened or how to prevent. All the connections were tight.
Specs: 4200W bulb running at 145amps 380/440v suffecient cooling.
The bulb still works from the switch underneath the lamphousing but the CNA200 automation 'lamphouse on/auto' switch is dead?
 -

 |  IP: Logged

Leo Enticknap
Film God

Posts: 7474
From: Loma Linda, CA
Registered: Jul 2000


 - posted 10-08-2004 02:59 AM      Profile for Leo Enticknap   Author's Homepage   Email Leo Enticknap   Send New Private Message       Edit/Delete Post 
Interesting... Darren came across a similar problem with a Cinemeccanica lamphouse at a student film society place he was servicing. The symptom was that the lamp either wouldn't strike at all or went out within a few minutes. The problem turned out to be a ceramic insulator plate against a bulkhead through which the anode lead passed. It wasn't totally gone, but had eroded to the point of allowing the lead to short against the bulkhead. I presume the cause was either heat damage or the insulator plate being a duff one to start with.

 |  IP: Logged

Stephen Furley
Film God

Posts: 3059
From: Coulsdon, Croydon, England
Registered: May 2002


 - posted 10-08-2004 04:57 AM      Profile for Stephen Furley   Email Stephen Furley   Send New Private Message       Edit/Delete Post 
quote: Tommie Evans
4200W bulb running at 145amps 380/440v
[Eek!]

I don't think that's what you mean, is it? That would be 55.1/63.8 kW! 380/440 is the mains Voltage (phase to phase), not the lamp Voltage. At least I hope it is, otherwise I think you might well vaporise a few things! [Smile]

 |  IP: Logged

John Pytlak
Film God

Posts: 9987
From: Rochester, NY 14650-1922
Registered: Jan 2000


 - posted 10-08-2004 05:58 AM      Profile for John Pytlak   Author's Homepage   Email John Pytlak   Send New Private Message       Edit/Delete Post 
Since the you say the connections were tight, and the wires themselves do not appear to be charred, I suspect that over time, the high voltage ignition had leaked across the insulating block (perhaps due to moisture or dirt), eventually burning it to the point that it began to conduct enough low voltage current to cause the severe damage you now see.

Any insulator for high voltage needs to be kept clean and moisture free, or there could be gradual breakdown, especially for insulators made of plastics. A ceramic insulator might have been a much better choice in this design.

 |  IP: Logged

Steve Kraus
Film God

Posts: 4094
From: Chicago, IL, USA
Registered: May 2000


 - posted 10-08-2004 03:39 PM      Profile for Steve Kraus     Send New Private Message       Edit/Delete Post 
Is there a way to run a line directly from ignitor to bulb without the need for an intermediate terminal?

 |  IP: Logged

Richard Fowler
Film God

Posts: 2392
From: Ft. Lauderdale, FL, USA
Registered: Jun 2001


 - posted 10-08-2004 03:50 PM      Profile for Richard Fowler   Email Richard Fowler   Send New Private Message       Edit/Delete Post 
This is common in older Optimax units: the U.V. from the bulb attacks the insulator and it will weaken. Heat also effects the insulator....nothing lasts forever [Wink] unless they desire to spend a few more $$$ in components. The route around the 45 degree mirror requires the mid way connection point.

 |  IP: Logged

Tommie Evans
Expert Film Handler

Posts: 116
From: Birmingham, West Midlands, UK
Registered: Sep 2004


 - posted 10-09-2004 05:55 AM      Profile for Tommie Evans   Email Tommie Evans   Send New Private Message       Edit/Delete Post 
quote:
I don't think that's what you mean, is it? That would be 55.1/63.8 kW! 380/440 is the mains Voltage (phase to phase), not the lamp Voltage. At least I hope it is, otherwise I think you might well vaporise a few things!
Sorry, that was the mains voltage - thanks for the advice the new part is on order [beer]

 |  IP: Logged



All times are Central (GMT -6:00)  
   Close Topic    Move Topic    Delete Topic    next oldest topic   next newest topic
 - Printer-friendly view of this topic
Hop To:



Powered by Infopop Corporation
UBB.classicTM 6.3.1.2

The Film-Tech Forums are designed for various members related to the cinema industry to express their opinions, viewpoints and testimonials on various products, services and events based upon speculation, personal knowledge and factual information through use, therefore all views represented here allow no liability upon the publishers of this web site and the owners of said views assume no liability for any ill will resulting from these postings. The posts made here are for educational as well as entertainment purposes and as such anyone viewing this portion of the website must accept these views as statements of the author of that opinion and agrees to release the authors from any and all liability.

© 1999-2020 Film-Tech Cinema Systems, LLC. All rights reserved.